RFID and Smart Packaging in healthcare
Case Study: RFID in Action - the
Massachusetts General Hospital START project
by Sunny Dzik, Co-Director Blood Transfusion Service, Massachusetts
General Hospital, USA
“To err is human”
In the year 2000, the Institute of Medicine published it’s landmark report,
“To err is human: Building a Safer Healthcare System”. This report detailed
current long-standing structural problems in the safe delivery of healthcare
to patients. Drawing largely on data concerning medication errors in
hospitals, the IOM report estimated that as many as 96,000 Americans may die
each year as the result of errors in the delivery of healthcare. The report
generated enormous attention to the problem of medical error and galvanized
governmental, regulatory, and professional groups to respond. Proper patient
identification—a critical requirement to give the right treatment to the
right patient-- has become widely understood as the fundamental
infrastructure of safe patient care. Organizations such as the Joint
Commission on the Accreditation of Hospitals (JCAHO) have clearly signaled
proper patient identification as a basic requirement of good healthcare and
a starting point for the assessment of any hospital. While at first glance
the identity of the patient seems a simple matter, current practice
demonstrates that much improvement is needed. We describe herein an example
of the problem of patient identification at the time of blood transfusion
and describe our initial steps to explore an improved method of patient
identification using RFID technology currently under study at Boston’s
Massachusetts General Hospital.
Blood Transfusion Errors
While the general public associates blood transfusion with the risk of
exposure to infectious hazards, transfusion of blood to the wrong patient (mis-transfusion)
is the most important serious hazard of transfusion.(1-4) The risk of mis-transfusion
is more than 100 times greater than the risk of HIV or HCV transmission by
blood.(1) Mis-transfusion typically results from an error made during the
bedside check just prior to transfusion.(1,3) Studies have documented that
such errors are most likely to occur among surgical patients.(3) Currently,
the bedside check is done by humans using eye-readable information in a
manner little changed in fifty years and without the advantage of any new
technology.
Data from multiple sources worldwide document that mis-transfusion errors
are serious and unacceptably frequent. In the United States, mis-transfusion
has consistently been the leading cause of death from transfusion reported
to the FDA since such reporting began. (3) Similar results are found in
other nations. The frequency of serious adverse events related to blood
transfusion is assessed in England through an organized national program of
hemovigilance called the Serious Hazards of Transfusion (SHOT) program.
Figure 1 categorizes all adverse transfusion events reported to the SHOT
program including not only those attributed to errors but also those
resulting from unavoidable blood reactions. When the data from SHOT are
further analyzed to only include those events that result in either death or
substantial morbidity to the recipient, mis-transfusion remains the most
important hazard. The most recent data from the SHOT program identified 213
incidents of incorrect blood transfused in the year 2002.(4) When analyzed
in detail, the most common cause of error was improper bedside check.
Figure 1: Serious Hazards of Transfusion (SHOT) national data from the UK(4)
Other sources report very similar results. Linden et al calculated that
the risk of ABO mis-transfusion was 1 in 12,000 units. (5) Pierre Robillard
reports that the hemovigilance program in Quebec identified mis-transfusion
as the most common major adverse event occurring at a rate of 1 in 12,000
transfusions.(6) Andreu reports similar findings from the hemovigilance
program in France.(7)
Current safety checks against mis-transfusion are
inadequate and getting worse
Current patient safeguards to prevent mis-transfusion are inadequate.
JCAHO has identified better patient identification at the time of blood
transfusion as a top priority for improved patient safety. Previous
published studies have repeatedly documented weaknesses in performance of
the final bedside clerical check.(9-11) Of particular concern is the
suggestion that the process is getting worse. The College of American
Pathologists recently published results of two large observational audits—
one conducted in 1994 and one in 2000.(12) These audits assessed the
frequency with which basic elements of the bedside check were performed
including positive patient identification, matching wristband identification
to the blood compatibility label, matching patient identifiers with the
blood request, and review of compatibility and expiration date information.
In each category, there was a substantial decline in the percentage of
transfusions that were correctly checked at the bedside between the years
1994 and 2000. Equally disturbing was the actual frequency that checks
failed to be performed. For example, in the year 2000, the audit of over
4,000 transfusions revealed a failure to match wristband identification with
the compatibility label in 25% of transfusions. Nationally, this translates
into literally millions of episodes in which patients are not provided basic
safeguards against the most common serious hazard of transfusion.
Technology support is appropriate for the bedside check:
Since most ABO incompatible transfusions (5) and most acute fatal blood
reactions (3) result from giving blood to the wrong recipient, the process
of the bedside check is the central target for improving patient safety
related to transfusion. The bedside check is a repetitive task subject to
“slip or lapse” type errors. (13) The essential features of the check are to
verify that the patient identifiers written on the compatibility label
attached to the blood bag match the identifiers of the actual patient about
to be transfused. Because different patients may have similar names and
medical record numbers, wrong units may be delivered to the bedside. The
final bedside check requires careful comparison of names and medical record
numbers of the patient and the blood label. Currently, the bedside check is
performed without the aid of any supportive technology-- even though
previous research has documented that the performance of repetitive task
functions is aided through the use of technology.(14)
Radio-frequency identification (RFID) for patient
and blood identification
Two machine-readable technologies—bar code technology and RFID—are
candidates for application to the bedside transfusion check. Bar code
technology has been available for decades, but has not gained acceptance for
blood checks at the bedside despite the fact that all blood containers have
been bar coded for many years. We believe the failure to adopt bar code
technology for the bedside transfusion check reflects its unsuitability for
bedside use. Bar coding requires line-of-sight so that a hand held laser can
read a flat surface with the code. The laser light must be oriented by the
user to cross the barcode. These simple constraints may represent important
practical obstacles to bar-coding especially in operating rooms where the
patient is covered with surgical drapes. Lack of acceptance of bar coding
suggests the need to develop other identification technologies.
START: Safer Transfusion with Advanced
Radiofrequency Technology
Massachusetts General Hospital has begun to study RFID “smart tag”
technology to improve the bedside check. Working with Precision Dynamics
Corporation, a major supplier of hospital wristbands and with Lattice
Corporation, a software company focused on improved patient safety, we have
begun to develop an RFID solution to the problem of transfusion error in the
Operating Room.
Radiofrequency technology
Passive smart tags can be engineered to use several different ranges of
radiofrequency from 125 kilohertz to 5.8 gigahertz. We propose to use tags
operating at 13.56 megahertz as these provide the following advantages: low
energy, low cost, no need for battery source in tag, small, will not
interfere with other medical equipment.
RFID technology is well-suited to rapid reading of the blood label in the
OR. RFID tags have been already employed in industrial applications
requiring fast read-rates such as checking products on conveyor lines and at
loading docks in supply chain management. Reading and writing data to the
RFID tags is fast and accurate, with the time required to transmit at less
than 100 milliseconds. Of particular importance for this proposal, data
transfer is automatic (does not require action by the user) as soon as the
tag enters the field of the reader device.
Making it “easy to do the right thing”
Correct reader design is critical to the success of the implementation
project. Failure to use the reader removes the patient-safety advantage
provided by the technology. In this application we propose an initial pilot
study under controlled conditions that will permit us to optimize our
existing reader design based on field experience, feedback, and data
obtained on user acceptance. Our goal is a reader design which is seamlessly
integrated into the process of administering a blood transfusion.
Why study this technology in the Operating Room?
We propose to pilot study the RFID check in the operating rooms (ORs).
Data from Sazama document that mis-transfusion events are most likely to
occur in surgical patients.(3) The OR is an ideal location for the
implementation of technology to assist in the pre-transfusion check for the
following reasons:
REASONS TO USE RFID BLOOD CHECKING IN THE OR
• A large proportion of the blood supply is transfused in the OR
• Patients are unconscious during the transfusion and cannot state their
name.
• Caregivers in the OR have limited contact-time with the patient and may
not “know” the patient as well as nurses on non-surgical floors.
• OR staff have experience with the use of technology designed for patient
safety.
• Blood is often given under urgent circumstances.
• Distraction in the OR is commonplace
At MGH, 25 % of red blood cell transfusions are administered in the ORs. No
other hospital location (patient floor, emergency room, outpatient area,
etc) transfuses more red blood cells than the OR. The opportunity for error
is particularly high as patients are unconscious and thus cannot state their
name at the time of the bedside check. (Asking the patient to state his name
is a standard part of the bedside check outside of the OR.) Unlike the nurse
outside the OR who may have cared for the patient over the course of several
days, nurses and anesthesiologists in the OR have not “known” the patient
for a prolonged period of time. The OR environment is “friendly” to the use
of technology to assist in patient safety. Indeed patient-safety devices
developed in conjunction with human factors research have been long applied
in anesthesia. (15,16) Finally, blood transfusions are often given in the OR
under circumstances of extreme urgency and distraction—two key elements that
may contribute to lapse errors at the time of the bedside check. Thus, from
both a perspective of need and acceptance we are confident that the OR is
the ideal hospital location to implement technology that assists in the
bedside transfusion check.
A simplified view of the system (a design showing reader input to LCD
monitor) is shown below:
Data Collection during the START study
The primary endpoints are user acceptance and system performance:
User acceptance
User acceptance will be measured as the proportion of units which were
checked by the users in the Operating Room with the RFID system. The
hospital blood bank data base records each unit issued, transfused, and
returned to the laboratory for each patient. The blood bank system also
records the date and time of issue and return, the type of product (RBC, FFP,
etc), the unit’s unique number, and the patient to whom the unit was issued.
By comparing blood bank records on each unit issued to the patient (and each
unit transfused) with the data recorded on the tablet computer attached to
the RFID reader in the OR reader, we can determine the proportion of issued
units (and proportion of transfused units) that were checked in the OR by
the RFID reader.
System performance
System performance will be measured by the ability of users to recognize
that the intended recipient of a labeled unit does not match the actual
recipient in the room. Because such errors are infrequent under routine
conditions, we propose to measure the performance of the system using “test
tag” units that act as a surrogate for mis-labeled units.
“Test Tag” units (surrogate mislabeled units) to test the system:
Test tag units will contain deliberately encoded information that does not
match the recipient and therefore act as surrogate mislabeled units. We have
developed a technique for applying these test tag units that will not
compromise patient safety during the evaluation.
Obstacles to RFID in the OR bedside check of
blood-- Future Opportunities for RFID
We envisage three potential obstacles to the success of RFID technology
applied to the bedside check of blood units in the operating room:
a. Failure to wrist-band the patient (this obstacle will evaporate if MGH
decides to apply RIFD wrist-bands to all patients);
b. Failure to initialize the reader; or
c. Failure to use the reader.
Our study will be able to measure with high precision the rates of each of
these three failure modes and permit us to analyze which part of the process
should be improved. For example, if failure to use the reader emerges as a
common obstacle, then “automatic readers” placed at the OR doorway would be
explored in a subsequent study.
Literature Cited:
1. Dzik WH. Emily Cooley Lecture 2002: Transfusion Safety in the Hospital.
Transfusion 2003; 43: 1190-99.
2. Minz PD. Nishot: on target, but there's no magic bullet. Am J Clin Pathol.
2001 Dec;116(6):802-5
3. Sazama K. Reports of 355 transfusion-associated deaths: 1976 through
1985. Transfusion 1990; 30: 583-590.
4. Love EM, Soldan K and the Serious Hazards of Transfusion Steering Group.
SHOT Annual Report 2000-2001, 2002, pp 1-239.
5. Linden JV, Wagner K,Voytovich AE, Sheehan J. Transfusion errors in New
York State: an analysis of 10 years’ experience. Transfusion 2000;40:
1207-1213.
6. Robillard P, Itaj NK, Corriveau P. ABO incompatible transfusions, acute
and delayed hemolytic transfusion reactions in Quebec hemovigilance system –
Year 2000 (abstract). Transfusion 2002; 42: 25S.
7. Andreu G, Morel P, Forestier F, et al. Hemovigilance network in France:
organization and analysis of immediate transfusion incident reports from
1994 to 1998. Transfusion 2002; 42: 1356- 64.
8. Baele PL, De Bruyere M, Deneys V, et al. Bedside transfusion errors. A
prospective survey by the Belgium SAnGUIS group. Vox Sang 1994; 66:117-21.
9. Shulman IA, Saxena S, Ramer L. Assessing blood administering practices.
Arch Pathol Lab Med 1999; 123: 595-598.
10. Zimmermann R, Linhardt C, Weisbach V, et al. An analysis of errors in
blood component transfusion records with regard to quality improvement of
data acquisition and to the performance of lookback and traceback
procedures. Transfusion 1999; 39: 351-356.
11. Miller KA. Transfusion errors. Q Probe09. Q-Probes 2000. College of
American Pathologists.
12. Novis DA, Miller KA, Howanitz PJ, Renner SW, Walsh MK. Audit of
transfusion procedures in 660 hospitals. Arch Pathol Lab Med 2003; 127:
541-48.
13. Leape LL. Error in medicine. JAMA 1994; 272: 1851-1857.
14. Bates DW, Cohen M, Leape LL, Overhage JM, Shabot MM, Sheridan T.
Reducing the frequency of errors in medicine using information technology. J
Am Med Inform Assoc 2001; 8: 299-308.
15. Cooper JB, Newbower RS, Lond CD, McPeek B. Preventable anesthesia
mishaps: a study of human factors. Qual Saf Health Care 2002; 11:277-82.
16. Cooper JB. Accidents and mishaps in anesthesia: how they occur, how to
prevent them. Minerva Anesthesiol 2001; 67: 310-3.
17. Dumont LJ. Sampling errors and the precision associated with counting
very low numbers of white cells in blood components. Transfusion 1999; 31:
428-32.
|
|
"I learned a lot, met people doing interesting
stuff, and now have to sort out where I want to go next! So much to
consider!!" Roz Ben-Chitrit, Avery Dennison
"Excellent. Well worth our time. Very pertinent to our needs"
Ray Rodgers, Maidstone & Tunbridge Wells NHS Trust
"Good mix of talk and demos" David Rodrick, NHSIA
"Informed, authoritative, and entertaining" Paul
Lafferty, Quintiles
|